Hi Carl,
I'm not sure what exactly you are asking. If you mean that if you change
the anode voltage from 400V to 500V and the gain goes from 2Alm to
10A/lm, will changing it from 500 to 600V give a gain of 50A/lm then the
answer is no the gain will be 30A/lm, you'd need to go to 700V to get
50A/lm. The graph "Anode Luminous Sensitivity and Gain Characteristics"
on the data sheet (fig 2 on my copy) shows this. It unfortunately a very
cluttered graph with non linear axes.
The other important question is what "mode" the PMT is being used in,
I.E single photon or current mode. If the light level is so low that you
are getting individual pulses (one for each photon) from the PMT then
adjusting the gain will not affect the number of pulses, just their
amplitude. If you are in current mode then the gain will directly affect
the signal level.
I do not know how the PCM2000 operates, but may instruments use a
current amplifier with a relatively long time constant. This when
individual pulses are present it integrates them and changes to current
mode when the out put becomes continuous. Have you tried imaging a
dilution series at different gain settings. Again I don't know the
PCM2000 but normally the "gain" control just adjusts the PMT anode
voltage. The other unknown is whether the ratio between the gain
"control" and the voltage is linear. Some instruments scale the voltage
change so that 10% change in the control really is 10% change in gain. A
dilution series check should show this.

HTH,
Robert.  

-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On
Behalf Of carl
Sent: 09 November 2004 00:04
To: [log in to unmask]
Subject: PMT sensitivity

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi All,
Does anyone have info on the linearity of the sensitivity of the
Hamamatsu
R928 PMT  that we have in our Nikon PCM2000 confocal?  Does it change
with
gain settings?  I have a 4-page spec sheet on this thing but I can't
decipher what it is telling me.
Thanks very much,
Carl

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-626-8469
FAX 520-621-3709
 
Any opinions expressed in this email are those of the individual and not necessarily Genetix Ltd (Genetix) or any company associated with it. This email and any files transmitted with it are confidential and solely for the use of the intended recipient. If you are not the intended recipient or the person responsible for delivering to the intended recipient, be advised that you have received this email in error and that any use is strictly prohibited. If you have received this email in error please notify Genetix by telephone on +44 (0)1425 624600. 
 

The unauthorised use, disclosure, copying or alteration of this message is strictly forbidden. This mail and any attachments have been scanned for viruses prior to leaving the Genetix network. Genetix will not be liable for direct, special, indirect or consequential damages as a result of any virus being passed on, or arising from alteration of the contents of this message by a third party. 
 