Subject: | |
From: | |
Reply To: | |
Date: | Mon, 12 Sep 2005 08:51:15 -0400 |
Content-Type: | text/plain |
Parts/Attachments: |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
At 07:19 AM 9/12/2005, you wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi all,
>
>I also have a very basic question on microscopy theory:
This is also a question that is deceptively basic!
>The Rayleigh criterion states that resolution d=0.61*lambda/NA. Taking the
>condenser lens into account this (in the textbooks) becomes d=
>1.22*lambda/(NAo+NAc). But shouldn't it be d= 0.61*lambda/NA with
>whichever NA is the smallest of the two?
Ok, first let me state that performance metrics like "Rayleigh criteria",
"Sparrow formula", "Modulation transfer function", "Point spread function",
etc. etc. etc. all have limited descriptive abilities and are not absolute
measures of the overall performance of an optical system. This is because
optics has a fundamentally incomplete conceptual foundation which I won't
get into here, other than to say Emil Wolf has done a monumental job of
partially reconciling- it has to do with describing how information is
carried in a beam of light.
So: the Rayleigh criteria was developed for imaging distant stars (point
objects) that are mutually incoherent- what happens to one star is
independent of what happens to the other. If I write your first formula as
d = 1.22*lambda/(2*NA), you can see that in the case of the condenser and
objective numerical aperture being equal, the two formulas agree. The
factor 1.22 is the first zero of a Bessel function, which is the
Fourier transform (Hankel, actually- the integration is performed in polar
coordinates) of a circle function.
Think of the formula as describing how the point spread function is carried
from one lens to the other. If one lens has a much lower NA than the
other, the point spread function is much larger, and any 'benefit' of the
high NA lens is lost: either one is subsampling a large PSF, or one is
undersampling a high PSF. Don't read too much into the formula- it's
valid only for low NA lenses anyway. Hope this helps!
If you want to discuss the further, email me directly,
Andy
Andrew Resnick, Ph. D.
Instructor
Department of Physiology and Biophysics
Case Western Reserve University
216-368-6899 (V)
216-368-4223 (F)
|
|
|