Hi, Stefan
Your assumption would be true if the light behaved symmetrically but in
actual fact, the condenser sets up the illumination for interaction with
the specimen, therefore we should consider its NA; then the light interacts
with the sample and all sorts of things happen (diffraction, refraction,
scatter, fluorescence, etc) that affect the angle of light exiting from the
sample, so we need also to include the NA of the objective, independently.
Another reminder: the NA written on the condenser is only the full NA;
closing the condenser aperture iris limits that to the real, working
NA. As a result, there is a big difference between the theoretically
possible R, derived from your equation, and the real, working R. Also, if
there is an iris in the objective, same thing.
Rule of thumb: The NA of the condenser should meet or exceed that of the
objective.
Hope that this is helpful.
Best regards,
Barbara Foster
Microscopy/Microscopy Education
313 S Jupiter Rd, Suite 100
Allen, TX 75002
P: 972-954-8011
W: www.MicroscopyEducation.com
P. S.
Need a good general reference or light microscopy text for the Fall? Call
us today to learn more about "Optimizing LIght Microscopy". Copies still
available through MME... even for class-room lots ... and we give quantity
discounts. Call Ken Piel at (972)954-8011.
At 06:19 AM 9/12/2005, Stefan Gunnarsson wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi all,
>
>I also have a very basic question on microscopy theory:
>
>The Rayleigh criterion states that resolution d=0.61*lambda/NA. Taking the
>condenser lens into account this (in the textbooks) becomes d=
>1.22*lambda/(NAo+NAc). But shouldn't it be d= 0.61*lambda/NA with
>whichever NA is the smallest of the two?
>
>tia
>Stefan
>
>+++++++++++++++++++++++++++++++++++++++++++++++++++++++