>I like the Hamamatsu ORCA AG for fixed and live cell fluorescence
>imaging. Very briefly...For live cell fluorescence imaging you
>want a cooled CCD camera that has high quantum efficiency (peak of
>60% or more in the wavelengths you will be imaging) and low read
>noise (8 electrons or less). If you are doing high resolution work,
>you want small pixels (7um or less is small enough for most of us).
>The capacity to do binning is also very useful for live cell
>imaging. I strongly recommend you do some reading as well. Take a
>look at: Spring (2000) BioTechniques Vol 21 (1):76. There are also
>good chapters on cameras in Digital Microscopy, 3rd Ed (Sluder &
>Wolf, eds). Appendix 3 in the 3rd edition of the "Confocal Bible"
>(Handbook of Biological Confocal Microscopy, Pawley, ed) is also a
>great resource. (Jim, haven't you made this available online?)
>
>Best, Jennifer
Hi all,
Yes, it can be found at:
http://www.springer.com/0-387-25921-X
Look under "Additional information" on the right and click the lower
of the 2 articles.
Cheers,
Jim P.
--
****************************************