Dear Gentlemen,
My name is Lukasz Kilianek.
I'm a Product Specialist at VisiTech International -manufacturer of VT
Infinity and VT Eye confocal units.
I follow your discussion about system comparison in term of speed and
photobleaching.
Let me point out that original mail was about VT Infinity. The last mail
pertains to VT Eye system.
Let me explain here - VT Infinity is a multiple point scanner, VY Eye is
a single spot scanner.
It makes a huge difference in performance.
With best regards
Lukasz Kilianek
Product Specialist
VisiTech International Ltd
Unit 92, Silverbriar
Sunderland Enterprise Park (East)
Sunderland SR5 2TQ, UK
Tel +44 191 516 6255
Fax +44 191 516 6258
email: [log in to unmask]
-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On
Behalf Of Martin Seem
Sent: 13 November 2007 13:57
To: [log in to unmask]
Subject: Re: Leica resonant scanner-live for cell imaging
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi Benedikt,
we acquired a Leica resonant scanner based confocal about 6 months ago.
Before buying this system we tested different types of microscopes
including three spinning disc based systems (Andor, PE and Visitec
QLC100), two point-scan based systems (Leica and Visitec Eye) and the
new Zeiss LSM5 Live system. We brought our own samples (transgenic plant
material) and used these to test the different systems in terms of
speed, photobleaching and overall image quality.
We found the spinning disc based systems to be slightly faster than the
Leica system especially when acquiring square images (e.g. 512x512
pixels). However the speed of the Leica system increases significantly
when you reduce the number of lines in the Y direction. You should have
no problem acquiring 10 fps with a resolution of 512x512 pixels
resolution and about 15-20 fps with a resolution of 512x128 pixels.
In our experience the level of photobleaching is quite low when using
the Leica system. I would say it is comparable to the spinning disc
systems we tested and significantly lower than the Visitec Eye system
that bleached my sample (transgenic GFP5-ER plants) within seconds (this
system did however work nicely for calcium imaging in dendrites so the
level of photobleaching depends on the properties of the fluorechrome).
In the end we chose the Leica system because it was sufficiently fast,
had an adjustable and round pinhole making it fully confocal, had low
levels of photobleaching, had a very flexible bandpass filter solution
and could easily and relatively cheaply be upgraded to a tandem scanner
system (both resonant and conventional scanner in the same system).
Best regards
Martin Seem
------------------------------------------------------------------------
-
Martin Seem
Graduate student
Group for Cell and Molecular Biology
Norwegian University of Science and Technology
|
|
|