Stephen Cody wrote:
>
> Presumably if you have CLEM fitted to your confocal you can choose to
> turn it off for quantitative measurements of fluorescence. Turning it on
> only when required for long time-lapse experiments where reducing
> bleaching is paramount, and where you are interested primarily in
> morphology rather than pixel to pixel intensity. Can someone please
> confirm it is implemented this way on the A1?
>
A couple of posts have noted that CLEM might interfere with intensity
measurement. I see no reason why this should be so. The CLEM system has all of
the information it needs to know how much exposure a given pixel received. That
being the case, the resulting intensity levels can be corrected accordingly.
Moreover, it would appear that CLEM operation allows one to avoid approaching
the edges of the detector's dynamic range, so that the detector can be operated
with a more linear response to intensity changes. Therefore, intensity
measurements should at least a accurate as a non-CLEM system.
Perhaps the developers can comment on whether my understanding is correct.
--aryeh
--