On Mar 3, 2010, at 8:09 AM, yee sarah wrote:
> Dear confocal listers,
>
> I've got a few questions about our Zeiss LSM510 system. And I am wondering if you could help me out with this.
>
> Under the scan control panel in the software, there are two methods to do the scanning : mean and sum.
> .......
>
> Another question is about the Line or Fram mode for averaging. People claim that fram averaging helps reduce photobleaching.
> If I use one direction scan for a fixed sample and my system is a laser point scanning confocal microscope.
Averaging or summing only reduces photobleaching if it allows you to avoid increasing the dwell time and/or laser intensity to achieve the same SNR.
Line operations may be fractionally faster that frame operations but both accomplish the same thing. Line averaging is usually preferred for live specimens since you will have less motion artifact between pixels.
>
> Would these two method apply the same photobleaching to my sample? In another word, would these two method have the same acquire time?
>
> Any feedback would be appreciated.
Regards,
Glen