Dear JP--
From your description, it might be lipofuscin. In my limited
experience, mouse tissue has a lot of it. If that's what it is, it will
appear yellow/orange with a wide-band UV filter, green through a
fluorescein filter, red through a rhodamine filter and will also be
visible through a Cy5 filter. It will be very resistant to
photobleaching. It will remain fluorescent after treatment with NaBH4
or oxidizing agents or after lipid extraction.
You can deal with lipofuscin one of at least 3 ways:
1) Don't use fluorescence--use immunoperoxidase methods.
2) Treat with Cu++. Cu++ ion will quench much of the autofluorescence
of lipofuscin, but it will also somewhat reduce the fluorescence of the
Cy3.
3) Treat with Sudan Black. Sudan Black binds lipophilic compartments
and, being black, quenches lipofuscin. It's not compatible with
xylene-based mounting media, though.
Steve Schnell, Bill Staines and I have a paper on this:
Schnell SA, Staines WA, Wessendorf MW. Reduction of lipofuscin-like
autofluorescence in fluorescently labeled tissue. J Histochem Cytochem.
1999 47(6):719-30.
Good luck!
Martin Wessendorf
On 4/5/2011 8:06 PM, Jean-Pierre CLAMME wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi,
>
> I'm looking for someone having experience with confocal imaging of liver
> tissue. I'm imaging mouse liver tissue and I have some issues with
> autofluorescence in the cy3 channel coming from structure appearing like
> vesicle in hepatocytes. Those vesicles mainly show up in the Cy3 channel
> but are also visible in the FITC channel (depending on the power I use).
> A spectral image with excitation at 488, shows a broad signal with a
> maximum around 580 nm.
> Could someone comment on the origin of this fluorescence ? Is it possible
> that it is lipofuscine ?
>
> Thank you,
>
> JP
--
Martin Wessendorf, Ph.D. office: (612) 626-0145
Assoc Prof, Dept Neuroscience lab: (612) 624-2991
University of Minnesota Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009
Minneapolis, MN 55455 e-mail: [log in to unmask]
|