Thank you very much !
Confocal Microscopy List <[log in to unmask]> wrote on
04/05/2011 06:35:56 PM:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
> Dear JP--
> From your description, it might be lipofuscin. In my limited
> experience, mouse tissue has a lot of it. If that's what it is, it will
> appear yellow/orange with a wide-band UV filter, green through a
> fluorescein filter, red through a rhodamine filter and will also be
> visible through a Cy5 filter. It will be very resistant to
> photobleaching. It will remain fluorescent after treatment with NaBH4
> or oxidizing agents or after lipid extraction.
> You can deal with lipofuscin one of at least 3 ways:
> 1) Don't use fluorescence--use immunoperoxidase methods.
> 2) Treat with Cu++. Cu++ ion will quench much of the autofluorescence
> of lipofuscin, but it will also somewhat reduce the fluorescence of the
> Cy3.
> 3) Treat with Sudan Black. Sudan Black binds lipophilic compartments
> and, being black, quenches lipofuscin. It's not compatible with
> xylene-based mounting media, though.
> Steve Schnell, Bill Staines and I have a paper on this:
> Schnell SA, Staines WA, Wessendorf MW. Reduction of lipofuscin-like
> autofluorescence in fluorescently labeled tissue. J Histochem Cytochem.
> 1999 47(6):719-30.
> Good luck!
> Martin Wessendorf
> On 4/5/2011 8:06 PM, Jean-Pierre CLAMME wrote:
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Hi,
> >
> > I'm looking for someone having experience with confocal imaging of
liver
> > tissue. I'm imaging mouse liver tissue and I have some issues with
> > autofluorescence in the cy3 channel coming from structure appearing
like
> > vesicle in hepatocytes. Those vesicles mainly show up in the Cy3
channel
> > but are also visible in the FITC channel (depending on the power I
use).
> > A spectral image with excitation at 488, shows a broad signal with a
> > maximum around 580 nm.
> > Could someone comment on the origin of this fluorescence ? Is it
possible
> > that it is lipofuscine ?
> >
> > Thank you,
> >
> > JP
>
> --
> Martin Wessendorf, Ph.D. office: (612) 626-0145
> Assoc Prof, Dept Neuroscience lab: (612) 624-2991
> University of Minnesota Preferred FAX: (612) 624-8118
> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009
> Minneapolis, MN 55455 e-mail: [log in to unmask]
|