CONFOCALMICROSCOPY Archives

April 2011

CONFOCALMICROSCOPY@LISTS.UMN.EDU

Options: Use Monospaced Font
Show Text Part by Default
Show All Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Subject:
From:
Craig Brideau <[log in to unmask]>
Reply To:
Confocal Microscopy List <[log in to unmask]>
Date:
Wed, 27 Apr 2011 20:18:43 -0600
Content-Type:
text/plain
Parts/Attachments:
text/plain (155 lines)
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

How much uncaging occurs outside the focal region in single photon?  If you
don't have your laser turned up to much the energy density should only
exceed the uncaging threshold near the focal point, yes?

Craig


On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[log in to unmask]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear all,
>
> Haven't followed this discussion closely, but would one advantage of 2P lie
> in the ability to uncage caged compounds - fluorescent tracers, for
> example,
> within a single cell?  We've used UV uncaging and while you get most
> uncaging in the target cell, you also get cones of uncaging above and below
> the plane of focus.  Seems that 2P would overcome this problem.
>
> Rosemary White
>
> Dr Rosemary White
> CSIRO Plant Industry
> GPO Box 1600
> Canberra, ACT 2601
> Australia
>
> T 61 2 6246 5475
> F 61 2 6246 5334
> E [log in to unmask]
>
>
>
> On 28/04/11 10:10 AM, "Craig Brideau" <[log in to unmask]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > I second the previous lister's opinions.  At the end of the day MP is
> really
> > only good for thick tissue sections.  If you are looking at cell layers
> just
> > use a conventional 1P confocal.  The key advantage of 2P is penetration
> > depth, which is moot when you are looking at cells.  That said, John's
> > comments about shorter UV imaging are still valid; it's easier to get a
> NIR
> > Ti:Saph beam through conventional optics than a very UV beam.  This only
> > applies if you are using dyes that need very UV excitation of course.
> >
> > Craig
> >
> >
> >
> >
> > On Sun, Apr 24, 2011 at 6:29 AM, George McNamara
> > <[log in to unmask]>wrote:
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >> Hi David,
> >>
> >> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic
> >> for exciting far red and infrared fluorophores. There were some
> exceptions,
> >> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755
> nm
> >> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since
> you
> >> are ok with cost and complexity, adding an OPO(s) and/or multiple MP
> lasers,
> >> gives you full range (see
> >> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for
> >> OPO).
> >>
> >> Depending on cell type and physiological needs, you may be able to
> suppress
> >> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of
> an
> >> artifact for many cell types - by using catalase/glucose oxidase/glucose
> or
> >> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and
> >> Mikhailov 1995 Cell Motil Cytokseleton).
> >>
> >> Enjoy,
> >>
> >> George
> >>
> >>
> >> On 4/23/2011 6:03 PM, David Knecht charter wrote:
> >>
> >>> *****
> >>> To join, leave or search the confocal microscopy listserv, go to:
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> *****
> >>>
> >>> Is there any disadvantage to multiphoton for cultured cells (besides
> cost
> >>> and complexity)? Cell viability?  Phototoxicity?  Dave
> >>>
> >>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote:
> >>>
> >>>
> >>>
> >>>> *****
> >>>> To join, leave or search the confocal microscopy listserv, go to:
> >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>>> *****
> >>>>
> >>>> Multiphoton has no clear advantage for cells in culture. For cell
> culture
> >>>> samples,
> >>>> we use two photon  only to excite DAPI or other UV and near-UV fluors
> >>>> since we
> >>>> had to make a choice between the Ti:S and a 405 laser on our scope.
> >>>>
> >>>> Kate Luby-Phelps
> >>>>
> >>>>
> >>> Dr. David Knecht
> >>> Department of Molecular and Cell Biology
> >>> Co-head Flow Cytometry and Confocal Microscopy Facility
> >>> U-3125
> >>> 91 N. Eagleville Rd.
> >>> University of Connecticut
> >>> Storrs, CT 06269
> >>> 860-486-2200
> >>> 860-486-4331 (fax)
> >>>
> >>>
> >>>
> >>
> >>
> >> --
> >>
> >>
> >> George McNamara, PhD
> >> Analytical Imaging Core Facility
> >> University of Miami
> >>
>

ATOM RSS1 RSS2