Hallo,
thanks everybody for their input regarding this topic. 
I have some results from a simple test.

Long dwell time (200 us) seem to produce essentially the same signal to 
background ratio  as accumulation of 100 images with 2us dwell times:
signal:  

1 x 200us dwell time:
                         area (pixels)	mean 	STDEV   signal-to-background
specimen             20812	             181.76	34.44     14.63
background          20812	              12.42	8.99

100 x 2 us dwell time, Sum of frames
                          area (pixels)	mean 	STDEV   signal-to-background
specimen              20812	             185.28	33.97     14.73
background           20812	              12.58	8.97


The specimen is a Chroma fluorescent slide occupying half of the field of view, 
the other half does not have any sample (= background).
The Sum was calculated in ImageJ (Image-Stack-Z Project-Sum of Slices).

FYI,
I also tested for the linearity of the signal versus laser power in the photon 
counring mode to see when the pile-up error would become noticeable.
Using 10 us dwell time, spectral detector was set to narrow width (2 nm band 
pass)to restrict the photon flux. The signal seems to scale well with intensity, 
with just a slight drop off at the highest laser powers (but some of that drop 
off can be due to photobleahching; after acquisition I zoomed out and the 
scanned area showed as a darker rectangle):
(results from an area of 16756 pixels)
                            Specimen             background
laser power (%)	mean 	STDEV      mean 	STDEV
0.1		9.36	7.62         0.67	2.13
0.2		11.3	8.42         0.65	2
0.3		13.26	8.89         0.65	2
0.4		15.21	9.69         0.67	2.1
0.5		16.85	10.22        0.68	2.09
1		45.54	16.63        0.68	2.13
2		82.51	22.64        0.67	2.15
4		155.69	31.58        0.66	2.09
8		297.19	43.26        0.73	2.19
16		567.71	60.79        0.79	2.32
32		1049.43	82.74        0.86	2.36


So it seems that I can count about 100 photons per microsecond without 
having to worry too much about the pile-up. Does anybody know how how the 
Olympus's "Hybrid photon counting" works? 

Spreadsheed file with more details, graphs are here:
http://microscopy.tamu.edu/instruments/light-microscopy/fv1000-tests

This folder also contains some of the image files.


Sincerely,

Stan Vitha