As I understand it, in principle there should be no difference.  But in
practice there could be a difference depending on the bit depths the
system could handle.  For example, early Bio-Rad systems (I'm showing my
age here) could only collect a single image at 8-bit depth, but averaged
into a 16-bit buffer, so you would do much better by averaging (so long
as you saved the final result as 16-bit).

                              Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon) 
Australian Centre for Microscopy & Microanalysis, 
Madsen Building F09, University of Sydney, NSW 2006 

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]]
On Behalf Of Stanislav Vitha
Sent: Friday, 17 June 2011 2:26 AM
To: [log in to unmask]
Subject: averaging vs. accumulation for noise reduction - is there a
difference?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hallo,
this is a very basic question, but I cannot figure this out from what I
have 
been reading, so a simple explanation for a non-physicist would be much 
appreciated:

Is there a real difference in the improvement of the signal to noise
ratio 
between frame averaging (or accumulation) and longer dwell times (slower

scan) for a point-scanning confocal witrh a PMT detector?

For instance, using single point scanning confocal, 12-bit acquisition.

a) averaging (or accumulating) 5 frames, 4 microseconds per pixel
b) acquiring a single frame, 20 microseconds per pixel

Assumptions: 
no saturation of the detector;
stable environmental conditions, no focus drift, etc

Would it matter (for the dfference between the two scenarios) if it was
analog 
detection or photon counting detection?

I will run this little test later, but I am curious what you think. 

I thought that at least for the photon counting mode, the two important 
factors would be the dark counts and the total number of counts
detected, so 
whether it is acquired in one scan or in 5 scans, it should be the same.
My 
camera expert here insists that the averaging scheme will give better
noise 
suppression.

Thanks!


Stan Vitha  
  