Low NA lenses reduce reflections by high incident angles and their long working distances provide tolerant focus. Just keep the focus spot less than the diameter of the detector area.
transmission efficiency can be estimated by focusing the measured objective and on an immersion objective on a clean cover slipped slide with Koehler illumination adjusted. then measure the mean intensity of a bright field image from each lens. You might have to place a ND filter to avoid saturation. If you want to be picky about wavelengths, perhaps place color filters in the light path.
I'm using a Thorlabs USB power meter, which can record time-lapse measurements with user-adjustable averaging.
Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Cellular Morphology Core
Center on Human Development and Disability
Box 357923
University of Washington
Seattle, WA 98195-7923 USA
(206) 616-4156
[log in to unmask]
[log in to unmask]
On Mar 12, 2013, at 10:04 AM, John Oreopoulos <[log in to unmask]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Julio, on the specifications tab of the weblink you sent, it says regarding objective compatibility:
>
> 4X-63X; air coupled, with FOV diameters less than 10mm
>
> Just to re-iterate, I don't think you can use an immersion objective with a power meter like this (someone from Lumen Dynamics can confirm this). Not only would it require you to place the optical sensor in contact with the immersion fluid of the objective (which might damage it), but even if you could do that, it still wouldn't work because these devices have a low acceptance angle for light, that is to say, these devices won't capture/register the light that approaches the sensor at very oblique angles which are typically present when using high NA immersion objectives.
>
> The absence of the immersion medium is also not an option because then many of the oblique rays will be lost at the front lens due to total internal reflection at the glass-air interface.
>
> John Oreopoulos
>
>
> On 2013-03-12, at 12:38 PM, Vazquez Lopez, Julio wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> You can check this:
>>
>> http://www.ldgi-xcite.com/products-xr2100-xp750.php
>>
>> We use a similar meter from Coherent, but i'm not in the lab now and can't tell you the model #. We routinely use power at the focal plane with up to 20x objectives. Not sure how accurate it would be with high NA… depends on how close you can get to the objective…
>>
>> Julio Vazquez.
>>
>>
>> ----- Original Message -----
>> From: "John Oreopoulos" <[log in to unmask]>
>> To: [log in to unmask]
>> Sent: Tuesday, March 12, 2013 9:09:35 AM
>> Subject: Re: Laser power meter and sensor
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi Martin, it regards to question 2, I don't think such a power meter exists (although if someone else knows otherwise, I'd love to know as well). You're better off measuring the power at the back of the objective (ie: unscrew the objective and measure the power at the empty nosepiece), and then use the wavelength transmission curve of the objective (usually available from the manufacturer) to calculate the percentage of light that gets through to the front end of the objective - some error is introduced here as there is some variability in the transmission for each identical objective, but at least you get a good estimate of the actual power on the other side of the objective.
>>
>> You might also consider over what area that power that gets to the other side of the objective is projected to determine the power density (power per unit area), which is is a more relevant parameter. There is a nice Nature Protocols paper from 2008 that talks about these kinds of measurements, and there is mention about choice of power meter I think:
>>
>> Grunwald, D., et al., Calibrating excitation light fluxes for quantitative light microscopy in cell biology. Nature Protocols, 2008. 3(11): p. 1809-1814.
>>
>> Cheers,
>>
>> John Oreopoulos
>> Staff Scientist
>> Spectral Applied Research
>> Richmond Hill, Ontario
>> Canada
>> www.spectral.ca
>>
>>
>> On 2013-03-12, at 11:17 AM, Martin Seem wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> A couple of questions for the list:
>>>
>>> 1, We are currently in the process of buying a new laser power meter that will
>>> primarily be used to check the general stability and power output of the laser
>>> lines of our CLSMs. Have anybody recently bought a power meter after
>>> considering some of currently offered products (like the Coherent LabMax
>>> TOP, Newport 1918-R, GigaHertz PT-9610 and Ophir Vega). Which one did
>>> you choose and why?
>>>
>>> 2, Are there any commercially available laser meter sensors that will allow us
>>> to measure the actual laser light output from a high NA (NA>1) water or oil
>>> immersion objective?
>>>
>>> Martin
|