CONFOCALMICROSCOPY Archives

August 2013

CONFOCALMICROSCOPY@LISTS.UMN.EDU

Options: Use Monospaced Font
Show Text Part by Default
Condense Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Mime-Version:
1.0
Sender:
Confocal Microscopy List <[log in to unmask]>
Subject:
From:
Alexia Ferrand <[log in to unmask]>
Date:
Fri, 23 Aug 2013 08:18:53 -0500
Content-Type:
text/plain; charset="ISO-8859-1"
Content-Transfer-Encoding:
quoted-printable
Reply-To:
Confocal Microscopy List <[log in to unmask]>
Parts/Attachments:
text/plain (23 lines)
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all,
We now figured out that the problem is coming from the temperature
controller (Oko-Lab). This periodicity of 66 sec seems to be the exact time
that the temperature/CO2 controller box takes to have the fan ON, then OFF,
then ON again. So every minute, the temperature is checked and if a
variation is detected, the chamber receives warm air (at about 48°C). And
somehow, this seems to influence the intensity/laser during the imaging. If
the controller box is turned off, the oscillations immediately stop. We
figured out that the user actually forgot a few times to fill in the water
dish containing the probe, and then the temperature controller was trying to
compensate. Once the probe was reimmerged and the system fully
re-equilibrated, I could not observe the oscillations anymore.
I contacted Oko-Lab while we were trying to figure out the problem, and they
have been very responsive. They pointed out a couple more things but the
probe not being in water still seems to be the main culprit.