CONFOCALMICROSCOPY Archives

November 2013

CONFOCALMICROSCOPY@LISTS.UMN.EDU

Options: Use Monospaced Font
Show Text Part by Default
Condense Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Sender:
Confocal Microscopy List <[log in to unmask]>
Date:
Sun, 17 Nov 2013 11:37:17 -0500
Reply-To:
Confocal Microscopy List <[log in to unmask]>
Content-Transfer-Encoding:
8bit
Subject:
From:
Lutz Schaefer <[log in to unmask]>
Content-Type:
text/plain; format=flowed; charset="UTF-8"; reply-type=original
In-Reply-To:
<2B5BFB55702DA54BA258D6FFC9289B5B62F3F687@ex-mbx-pro-06>
Organization:
AIMC
MIME-Version:
1.0
Parts/Attachments:
text/plain (158 lines)
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Guy,
in any incoherent modality, the image formation is always composed out of 
the product of emission and excitation psf's. In the special case of point 
scanning confocal microscopy, its only that the (widefield) emission psf is 
further convolved by the geometry of the pinhole before multiplication. 
Therefore, of course you are right in saying that towards a more open 
pinhole the emission psf component approaches a constant. In 2p, the 
excitation psf is significantly smaller (approaching the square with no 
aberrations present) and in theory a pinhole in the emission path could 
further increase the sectioning resolution. However, numerically this is 
insignificant as due to the square, the excitation psf already is much 
smaller than the emission psf could possibly become, even with pinhole. On 
top of all the often very low 2p SNR and present SA forbids to observe 
minute improvements.

Regards
Lutz

-----Original Message----- 
From: Guy Cox
Sent: Sunday, November 17, 2013 1:18 AM
To: [log in to unmask]
Subject: Re: 2P vs 1P psf

The confocal psf is not the product of excitation and emission psfs unless 
the pinhole is very small (much less than 1 Airy unit).  In normal use 
(pinhole at 1 Airy) it is just the excitation psf.  So it will actually be 
quite similar to the 2p one.  (At 800nm Rayleigh resolution is ~350nm and 
350/1.4 gives ~250nm.)

I agree with an earlier correspondent that the greater background is almost 
certainly due to the fact that 2p excitation spectra tend to be broader than 
1p ones.  If so it should be possible eliminate it by a more selective 
detection filter.

                                            Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On 
Behalf Of Jeff Reece
Sent: Sunday, 17 November 2013 1:13 AM
To: [log in to unmask]
Subject: Re: 2P vs 1P psf

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

First I will assume that "all other things equal" means that the 2P and 1P 
images have the same SNR.  (A lower SNR will give you a lower effective 
resolution.)

If you are within a few microns of the cover glass, resolution should be 
worse with 2P than 1P.  This is because the much longer excitation 
wavelength in 2P creates a larger psf.  Even though the effective psf in 2P 
mode is the square of the raw excitation psf, the confocal psf is the 
product of the excitation and emissions psfs.

Then, as you get away from the cover glass, at some point-- the resolution 
in the 2P image will start to look better than 1P.  This is because the 
shorter excitation wavelength in 1P has a harder time focusing at the deeper 
depths, due to the difference between immersion media RI and sample RI, 
and/or the dispersion of the sample as a function of wavelength.  Exactly 
where the transition occurs depends on these factors, maybe others that have 
escaped me at the moment.  So I say "a few microns" but it is more likely in 
the tens of microns when using water immersion with live sample.

Since your user is in the realm where 1P looks better then 2P, then it is 
likely that there is also not much dispersion of the emission through the 
sample, so one could collect the 2P emission on the descanned detector with 
the pinhole closed down (to a little less than 1AU based on the 2P psf), 
without losing much signal from the focal plane.  Resolution in 2P could 
then be increased, with the added benefit of removing other background from 
OOF planes.  Of course this doesn't work in thick tissue where there is 
dispersion.

If the 2P "background" signal you mention is not due to nearby OOF signal, 
but seems ever-present and non-changing as you move around a sparsely 
stained sample in xy:
2P requires a lot more laser power at the sample than 1P to get the same 
signal back; the extra power density in the focal volume is necessary in 
order for 2P to
occur.  Thus reflection of the laser on all optical interfaces, relative to 
the fluorescence signal, is much higher with 2P than 1P at the point where 
the emission hits the dichroic.  So the dichroic and emission filter need to 
be better at blocking the laser for 2P than 1P.
If you have this problem, then obviously the best fix is to replace or 
double up the emission filter so you get better blocking of the laser.  If 
your user is in the ("thin sample") case mentioned above, then you could 
alternatively use the quick fix of sending the 2P emission to the descanned 
detector and closing down the pinhole.

I hope my lengthy explanation has helped.

Cheers,
Jeff




>________________________________
> From: "Laevsky, Gary S." <[log in to unmask]>
>To: [log in to unmask]
>Sent: Friday, November 15, 2013 9:00 AM
>Subject: 2P vs 1P psf
>
>
>*****
>To join, leave or search
the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>Hi All,
>
>Iā€™m sure there is going to be a simple answer here, but it alludes me.
>
>I know the psf is a function of NA and wavelength.
>
>Obviously, with 1P we use a pinhole to exclude out of focus light, and when 
>set at ā€œ1 Airy,ā€ you have maximized the pinhole for the particular 
>wavelength you are using.  With 2P, no pinhole is necessary because of the 
>non-linear excitation mechanism.
>
>A user just approached me and asked if/why there would be more out of focus 
>light in a 2P image then in a 1P image with a properly set pinhole.  In a 
>collaborators experiments, there seems to be more background in the 2P 
>image (all other things equal).
>
>Only thing I can think of is a poor beam
profile on the 2P.  Maybe a large pulse width would excite a larger spot?
>
>Thankful for the insight.
>
>
>Best,
>
>Gary
>
>
>
>Gary Laevsky, Ph.D.
>Confocal Imaging Facility Manager
>Dept. of Molecular Biology
>Washington Rd.
>Princeton University
>Princeton, New Jersey, 08544-1014
>(O) 609 258 5432
>(C) 508 507 1310
>
> 