To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****
Dear Oliver,
I was also confused by the literature on Evans Blue fluorescence. Strangely, the best combination for the optics on a good old MRC-1024 was to use an excitation of 568 nm and a 585 nm dichroic. This allowed visualizing motoneurons retrogradely labeled with Evans Blue that was previously injected in the muscle. See the following papers for some images:
Nguyen et al. (2004) J. Physiol. 556, 203-219
Nguyen and Kleinfeld (2005) Neuron 45, 447-457
Maybe the fluorescence depends on where Evans Blue is bound. Also, I could not get any two-photon images with Evans Blue with the same specimens, owing possibly to the symmetrical structure of Evans Blue that would preclude multiphoton excitation.
Hope this helps.
Quoc Thang Nguyen
[log in to unmask]
On Nov 5, 2013, at 12:08 PM, Oliver Biehlmaier wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Thanks for the links and answers! That helps.
> We might also go for measuring with a spectrophotometer...
> Cheers,
> Oliver
>
> ----------------------------------------------------------------
>
> Oliver Biehlmaier, PhD
>
> Head of Imaging Core Facility
>
> Biozentrum (Kragenbau, Room G1054)
> University of Basel
> Klingelbergstrasse 50/70
> 4056 Basel
> Switzerland
>
> Office: +41 (61) 267 20 73
> Lab: +41 (61) 267 22 50
> Email: [log in to unmask]
> http://www.biozentrum.unibas.ch/imcf
> http://microscopynetwork.unibas.ch
> ----------------------------------------------------------------
>
>
>
>
> Am 05.11.2013 um 17:30 schrieb Chris Tully:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Oliver,
>>
>> I would highly recommend asking around among your local colleagues for someone who has a fluorescence spectrophotometer, and measure the curves yourself.
>>
>> Alternatively, I have read this article because I do not have access to it at the moment but according to Google it does have a UV-Vis absorbance spectrum.
>>
>> Chris
>>
>> On 11/5/2013 10:45 AM, Oliver Biehlmaier wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Dear listers,
>>>
>>> I have a question regarding Evans Blue:
>>>
>>> I have a user that is doing standard 3ch-labelings on a confocal microscope (DAPI, Alexa488, Alexa 568). Now she is having some sections from mice that have been injected with Evans Blue.
>>> When imaging these new samples, she wants to use the same 3ch approach as before but she does not want to see any fluorescence from the Evans Blue.
>>> I tried to find out what the excitation and emission spectrum of Evans Blue is but I could not find any clear information. E.g. one publication specifies the spectrum of Evans Blue as follows: excitation at 620 nm, emission at 680 nm (Saria A, Lundberg JM. Evans blue fluorescence: quantitative and morphological evaluation of vascular permeability in animal tissues. J Neurosci Methods. 1983 May;8(1):41-9. PubMed PMID: 6876872.) Others claim that it fluoresces with excitation peaks at 470 and 540 nm and that it has its emission peak at 680 nm. I tend towards the first spectrum but I did not find any other confirmation for it.
>>>
>>> Does anyone of you guys know the exact spectrum of that dye?
>>> This would be great!
>>>
>>> Cheers,
>>> Oliver
>>>
>>> ----------------------------------------------------------------
>>>
>>> Oliver Biehlmaier, PhD
>>>
>>> Head of Imaging Core Facility
>>>
>>> Biozentrum (Kragenbau, Room G1054)
>>> University of Basel
>>> Klingelbergstrasse 50/70
>>> 4056 Basel
>>> Switzerland
>>>
>>> Office: +41 (61) 267 20 73
>>> Lab: +41 (61) 267 22 50
>>> Email: [log in to unmask]
>>> http://www.biozentrum.unibas.ch/imcf
>>> http://microscopynetwork.unibas.ch
>>> ----------------------------------------------------------------
>>
>>
>> --
>> *Chris Tully*
>> Principal Consultant
>> Image Incyte, LLC
>> 240-475-9753
>>
>>
>> Image Incyte, LLC
>>
>> <http:%5C%5Cwww.ImageIncyte.com> [log in to unmask] <mailto:[log in to unmask]>
|