Hi Long,
Is it Coherent CUBE? You may ask Coherent for a quote. You'll find its
probably way less than half of the Leica price. But From Leica you get all
the service and guarantee and minimum downtime.
For a microscopy facility I'd go for Leica solution.
If it's my own microscope I'd consider bypassing Leica. But remember that
you need additional optical fiber, laser coupling and probably some parts
inside the scanhead. And don't forget the laser control and changes in the
software settings. So maybe the official way is the only solution for you.
Replacing the laser is so much simpler than adding a new one...
Good luck,
zdenek svindrych
---------- Původní zpráva ----------
Od: [log in to unmask] <[log in to unmask]>
Komu: [log in to unmask]
Datum: 5. 3. 2014 21:16:11
Předmět: price for adding 405nm laser in Leica SP5
"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****
Hi Everyone,
Right now we are planning to add 405nm laser line to our Leica SP5 II
confocal system. I got a quote from the Leica rep but would like to know if
that is the best price. I wonder if anyone can share your experience/quote
with me. Thanks.
Long"