Subject: | |
From: | |
Reply To: | |
Date: | Fri, 25 Aug 1995 13:57:59 +0100 |
Content-Type: | text/plain |
Parts/Attachments: |
|
|
On Thu, 24 Aug 1995, Guy Cox wrote:
> The beam diameter, assuming that your specimen is in a mountant
> of refactive index 1.515, will be an Airy disk at the point of focus,
> since the laser source should approximate to a point.
>
> Thus the radius (centre to 1st minimum) will be
> 0.61 * lambda / NA = 0.61 * 488 / 1.4 = 213nm
> for the 488nm line.
also assuming the BFP of the lens is 100% filled.
Regards,
******************************************************************************
Stamatis Pagakis
University Laboratory of Physiology [log in to unmask]
University of Oxford, Parks Rd. FAX: +44 (0)1865 272469
Oxford OX1 3PT, United Kingdom Tel: +44 (0)1865 272551
|
|
|