There is a measured c o n f o c a l spot in
Appl. Phys. Lett, 64, No. 11, 1994, p. 1335-1337 (Hell, et al).
The spot is measured for NA=1.4, (Zeiss 100X, oil Planapochromat) 633 nm.
The lateral full-width-at-half-maximum (FWHM) was 200 nm, the axial FWHM
was 520nm. The extent of the PSF scales with the wavelength.
The PSF measurement was carried out for an index of refraction of the
specimen matched to the oil immersion system. A mismatched refractive index
of the sample enlarges the spot (especially) in axial direction. This
enlargement depends on the focusing depth. For further reference you can
look up the 'Handbook' or one of the recent issues of 'Bioimaging'.