Volume calculations are relatively straightforward with the confocal
microscope when an excluded dye (e.g. fluorescein dextran) is introduced to
the media. We published a technique for making these calculations from 3D
reconstructions. The technique is fairly precise, although it is difficult
to assess its accuracy. However, this method requires a volume set of at
least 15-20 slices. Depending on the speed of your microscope, it may not
be fast enough for your needs. The articles are:
Guilak, F. Volume and surface area measurement of viable chondrocytes in
situ using geometric modelling of serial confocal sections. J Microsc
173:245-256, 1994.
Guilak, F. Ratcliffe, A. Mow VC. Chondrocyte deformation and local tissue
strain in articular cartilage: A confocal microscopy study. J Orthop Res
13:410-421, 1995.
Farshid Guilak
[log in to unmask]
---------------------
>Hi,
>
>We would like to measure changes in the volume of cultured mammalian
>cells. The changes occur in about 10-20 seconds, and we would like to
>do some kinetics studies on them. So we need fairly fast sampling rate
>(like every second or half-second). We have a ZEISS LSM that can be
>used for the experiments.
>
>Any advice, including pointers to publications on the subject would
>be greatly appreciated. Thanks!
>
>Laszlo Menczel
>Biological Research Center
>Szeged, Hungary
>
>[log in to unmask]
|
|
|