Following on from what others have said, it probably depends on how MetaMorph is reading the camera out (and what it is doing to the data afterwards - e.g. converting to photon counts etc ...). http://www.andor.com/learning-academy/dual-amplifier-dynamic-range-scmos-dynamic-range has some interesting info on how the Neo get's it's large dynamic range. Notable on that website is the following sentence "Due to the splicing together of the low and high gains, the transition region between them is not seamless but has been optimised as far as possible." To me this seems to suggest that if you're going to do deconvolution on the data (or any other form of quantitative data analysis), you might want to choose either the high or the low gain 11 bit mode, and live with the reduced dynamic range. It's quite possible that the MetaMorph engineers have made this choice when writing their camera code (It's certainly the safe default choice as it'll be sure not to
 generate artefacts whereas the combined mode might - knowing for sure would require a lot of testing, which might well be application specific).

best wishes,
David

________________________________
From: Lu Lei <[log in to unmask]>
To: [log in to unmask] 
Sent: Friday, 26 July 2013 3:56 AM
Subject: Neo scMOS problem


*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,
We recently purchased a wide-field microscope system with a scMOS camera 
(Neo, 5.5 Megapixel, by Andor). We are having quite some annoying problems 
with this Neo scMOS camera. Though it is a 16-bit device, it does not have the 
dynamic range it should have in our system.
1)    In “Rolling” shutter mode, the intensity value that the camera could 
generate is truncated at ~20K. In another word, the camera saturates at ~20K 
instead of 65535.
2)    In “Global” shutter mode, the camera saturates at ~57K—a little short of 
65535.
The camera is controled by the latest MetaMorph. The cooling temperature of 
the camera seems also normal as displayed by MetaMorph (-30C). We haven’t 
received any explanation from Andor via local vendor yet. I am wondering what 
is configured incorrectly here. 
Does anyone have similar experience of this camera or suggestion to share?
Thanks a lot.
Regards,
Lei Lu
-------------------------------------------------------------------------