Just to follow on with what Mark said, if all you want is to measure
distortion, you don't even need a reflection image of your grid. Just
use a transmission detector and, if you don't have one (assuming
inverted optics), a PIN diode sitting on the top of the slide should
do nicely, as TPE usually has a massive (mW) beam.
JP
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>Post images on http://www.imgur.com and include the link in your posting.
>*****
>
>Unfortunately, it is extremely difficult to record the reflection mode
>image in our microscope, as the NIR light will not pass through the
>dichroic, and even if we changed out the dichroic for a beam splitter,
>out PMTs are not sensitive to NIR light.
>
>Translating a bead though is a very interesting idea. We use a high
>precision thorlabs MLS stage. I will double check the repeatability,
>but in theory I could program it to perform an extremely high
>resolution distortion measurement. Of course, getting a single bead
>onto a coverslip (or else programming software to automatically
>identify the same bead across hundreds of frames) will be somewhat
>tricky.
>
>Its a shame that getting targets for multiphoton is so difficult.
>Ideal I suppose would be something etched in glass so that I could
>flow my own fluorophore through it. I will have to look around and
>see if I can get a lithography facility to make something like that.
>
>Thanks,
>Mike
>
>On Wed, Jul 23, 2014 at 3:19 AM, Mark Cannell
><[log in to unmask]> wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Thinking about this a bit more, I realised that your existing grid
>>is perfectly suitable as you don't need 2P excitation to measure
>>field distortion at the wavelength you are using, Just turn the
>>power down a lot and record a reflection image,
>>
>> Cheers Mark
>>
>>>>
>>>> Hi Mark,
>>>>
>>>> I'm assuming you mean beads deposited in a grid or something similar?
>>>> Could you point me to where you found this?
>>>>
>>>> Thanks,
>>>> Mike
>>>>
>>>> On Tue, Jul 22, 2014 at 3:58 AM, Mark Cannell
>>>> <[log in to unmask]> wrote:
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> Post images on http://www.imgur.com and include the link in your
>>>> posting.
>>>>> *****
>>>>>
>>>>> Hi Michael
>>>>>
>>>>> I've always used beads to test the microscope. They are relatively cheap
>>>> and if you blow one up it don't matter. As you say, evaporated metal
>>>> targets are too easily damaged (as Jim Pauley and I discovered
>>>>when testing
>>>> my microscope back in '95!)
>>>>>
>>>>> HTH
>>>>>
>>>>> Mark
>>>>>
>>>>> On 22/07/2014, at 4:30 am, Michael Giacomelli <[log in to unmask]> wrote:
>>>>>
>>>>>> *****
>>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>>> Post images on http://www.imgur.com and include the link in your
>>>> posting.
>>>>>> *****
>>>>>>
>>>>>> Could someone recommend a good multiphoton distortion target?
>>>>>> Preferably one that is relatively robust against damage.
>>>>>>
>>>>>> Currently I am using a thorlabs wire grid target with fluorophore
>>>>>> behind it. The beam is attenuated by the metal, resulting in an image
>>>>>> of the grid. However, passing through the entire 1 mm slide greatly
>>>>>> reduces my resolution, and unless I am extremely careful, the grid is
>>>>>> ablated by the beam.
>>>>>>
>>>>>> Thanks,
>>>>>> Michael
>>>>>
>>>>>
>>>>> Mark B. Cannell Ph.D. FRSNZ
>>>>> Professor of Cardiac Cell Biology
>>>>> School of Physiology & Pharmacology
>>>>> Medical Sciences Building
>>>>> University of Bristol
>>>>> Bristol
>>>>> BS8 1TD UK
>>>>>
>>>>> [log in to unmask]
> >>>
>>
>>
>>
>> Mark B. Cannell Ph.D. FRSNZ
>> Professor of Cardiac Cell Biology
>> School of Physiology & Pharmacology
>> Medical Sciences Building
>> University of Bristol
>> Bristol
>> BS8 1TD UK
>>
>> [log in to unmask]
--
Prof. James Pawley, Shangri La Hotel, Harbin,
email <[log in to unmask]>, China Cell phone 15636810175
|