Hello, everyone,
Just like to emphasize one thing that others already mentioned: The
photobleaching rate is proportional to the SQUARE of the zoom-factor.
Therefore, increasing zoom unnecessarily (beyong Nyquist sampling rate)
will bleaching the sample much faster. Therefore, you may even get a
worse (bleached) picture.
Xuejun
Guy Cox wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Gary wrote:
>
>
>>This question has just been posed and I need a correct answer.
>>
>>I know that zooming on our Radiance 2000 when using a 63x oil
>>objective anymore than 2 to 3 times is a "waste." The precise
>>definition of "waste" is what I'm looking for. I have been told that
>>this will yield a digital artifact, due to artificial "filling," as a
>>result of decreased resolution.
>>
>>
>
>
>You have been given somewhat inaccurate information. 'Zooming'
>on a confocal is just scanning a smaller area, so there is no
>digital filling involved. The key question you need to ask is
>how many pixels you need to actually get the full resolution
>your lens can give you - with this lens 200nm or better.
>
>The Nyquist criterion says that you need a minimum of 2.3 pixels
>within the minimum resolved distance to be able to capture that
>resolution in a digital image. At zoom 3 your Radiance has a pixel
>size of 120nm so it is clear that you will NOT use the full resolution
>of your lens at zoom 3 or below. At zoom 4 you are getting fairly
>close to Nyquist (pixel size 90nm).
>
>I believe that a _small_ amount of oversampling is a good thing
>(your lens should do a bit better than 200nm and/or you may want
>to smooth the image a bit to reduce noise, etc. etc) so I'd go to
>a maximum of 3 pixels per minimum resolved distance. This isn't
>enough to make the image look fuzzy - in fact for visual presentation
>it looks pretty well perfect. This takes you somewhere between
>zoom 5 (pixel size 70nm) and zoom 6 (pixel size 60nm).
>
>Anything above zoom 6 will start to look fuzzy since you are in
>'empty magnification' - there just isn't any fine detail to see
>and your eye doesn't like that. Also you will bleach your sample
>more - to no purpose. So make zoom 6 the limit with this lens.
>
>All these figures are based on 512 x 512 image - obviously this
>will be different for different image resolutions. But the Bio-Rad
>always shows you the actual pixel size so it's easy to see if you
>are in the right area.
>
> Guy
>
>
>Assoc. Prof. Guy Cox, ooOOOOOOoo
>E.M. Unit, F09 # oOOOO | | OOOOo #
>University of Sydney ### OOO| | | | | |OOO ###
>NSW 2006, Australia ### OOO | | | | | | OOO ###
>Ph: 02 9351 3176 ### OO | | | | | | | | OO ###
>Fax: 02 9351 7682 ##### | | | | | | | | #####
> ==#####============================#####==
>http://www.guycox.net ##### #####
>http://www.guycox.com ~~#####~~~~~~~~~~~~~~~~~~~~~~~~~~~~#####~~
>
>
>
--
Xuejun Sun, Ph.D.
Dept. Exp. Oncology
Cross Cancer Institute
11560 University ave.
Edmonton Alberta T6G 1Z2
Canada
Phone: (780) 432-8898 (office)
(780) 432-8468 (lab.)
Fax: (780) 432-8425
This e-mail and any attachments may contain confidential and
privileged information. If you are not the intended recipient,
please notify the sender immediately by return e-mail, delete this
e-mail and destroy any copies. Any dissemination or use of this
information by a person other than the intended recipient is
unauthorized and may be illegal.
|
|
|