Hi Carl,
Mark and Guy have already answered this and I'd agree with what they
say. Basically once the system is stabilised at a reasonable gain
setting the PMT will be linear to much less than 1 count of your 8 bit A
to D. Assuming the amplifiers are reasonable you should have no problem.
On the subject of variation between tubes, we have our PMT's specially
selected by the manufacturer for better matching. 
One other factor that can cause problems is if the PMT has been exposed
to excessive light levels. This causes problems even if the tube is not
powered at the time. A PMT can take many hours to recover from even a
few seconds of exposure to room lighting.

Regards,
Robert.

-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On
Behalf Of carl
Sent: 09 November 2004 17:26
To: [log in to unmask]
Subject: Re: PMT sensitivity

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Robert,
Thanks very much for the reply.  However my background doesn't include
the
wherewithall to gleen from your reply the answer to my question.

I will rephrase: At a given gain setting, do pixel values from 0-255 in
an
image imply a linear increase in brightness in the sample?  If not, how
does
one correct for the non-linearity?  When the gain is changed to a
different
setting, is there a change in the slope/shape of the sensitivity curve?
The
user that prompted this question wants to measure greyscale values and
wants
to know if a range of low values have the same relationship to each
other as
mid-values have to each other and high pixel values have to each other.
I
guess the next question is, "Is there a simple answer to this question?"

You identified several unknowns regarding how the PMT is being used in
the
system.  These questions could probably be best be answered by the Nikon
folks who made the system, however I've received no response yet from a
direct query.   Is anyone out there?

Thanks again,
Carl
Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-626-8469
FAX 520-621-3709
----- Original Message -----
From: "Robert Atkinson" <[log in to unmask]>
To: <[log in to unmask]>
Sent: Tuesday, November 09, 2004 3:59 AM
Subject: Re: PMT sensitivity


> Hi Carl,
> I'm not sure what exactly you are asking. If you mean that if you
change
> the anode voltage from 400V to 500V and the gain goes from 2Alm to
> 10A/lm, will changing it from 500 to 600V give a gain of 50A/lm then
the
> answer is no the gain will be 30A/lm, you'd need to go to 700V to get
> 50A/lm. The graph "Anode Luminous Sensitivity and Gain
Characteristics"
> on the data sheet (fig 2 on my copy) shows this. It unfortunately a
very
> cluttered graph with non linear axes.
> The other important question is what "mode" the PMT is being used in,
> I.E single photon or current mode. If the light level is so low that
you
> are getting individual pulses (one for each photon) from the PMT then
> adjusting the gain will not affect the number of pulses, just their
> amplitude. If you are in current mode then the gain will directly
affect
> the signal level.
> I do not know how the PCM2000 operates, but may instruments use a
> current amplifier with a relatively long time constant. This when
> individual pulses are present it integrates them and changes to
current
> mode when the out put becomes continuous. Have you tried imaging a
> dilution series at different gain settings. Again I don't know the
> PCM2000 but normally the "gain" control just adjusts the PMT anode
> voltage. The other unknown is whether the ratio between the gain
> "control" and the voltage is linear. Some instruments scale the
voltage
> change so that 10% change in the control really is 10% change in gain.
A
> dilution series check should show this.
>
> HTH,
> Robert.
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[log in to unmask]]
On
> Behalf Of carl
> Sent: 09 November 2004 00:04
> To: [log in to unmask]
> Subject: PMT sensitivity
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi All,
> Does anyone have info on the linearity of the sensitivity of the
> Hamamatsu
> R928 PMT  that we have in our Nikon PCM2000 confocal?  Does it change
> with
> gain settings?  I have a 4-page spec sheet on this thing but I can't
> decipher what it is telling me.
> Thanks very much,
> Carl
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> University of Arizona
> 520-626-8469
> FAX 520-621-3709
>
> Any opinions expressed in this email are those of the individual and
not
> necessarily Genetix Ltd (Genetix) or any company associated with it.
This
> email and any files transmitted with it are confidential and solely
for
> the use of the intended recipient. If you are not the intended
recipient
> or the person responsible for delivering to the intended recipient, be
> advised that you have received this email in error and that any use is
> strictly prohibited. If you have received this email in error please
> notify Genetix by telephone on +44 (0)1425 624600.
>
>
> The unauthorised use, disclosure, copying or alteration of this
message is
> strictly forbidden. This mail and any attachments have been scanned
for
> viruses prior to leaving the Genetix network. Genetix will not be
liable
> for direct, special, indirect or consequential damages as a result of
any
> virus being passed on, or arising from alteration of the contents of
this
> message by a third party.
>
>
 
Any opinions expressed in this email are those of the individual and not necessarily Genetix Ltd (Genetix) or any company associated with it. This email and any files transmitted with it are confidential and solely for the use of the intended recipient. If you are not the intended recipient or the person responsible for delivering to the intended recipient, be advised that you have received this email in error and that any use is strictly prohibited. If you have received this email in error please notify Genetix by telephone on +44 (0)1425 624600. 
 

The unauthorised use, disclosure, copying or alteration of this message is strictly forbidden. This mail and any attachments have been scanned for viruses prior to leaving the Genetix network. Genetix will not be liable for direct, special, indirect or consequential damages as a result of any virus being passed on, or arising from alteration of the contents of this message by a third party. 
 

ATOM RSS1 RSS2