Dear Nathalia,
We had a problem like this on our SP2 some years ago. This may not be that relevant because we have an upright scope with a galvo stage but it turned out that the galvo actually needed recalibrating. We had a Leica engineer recalibrate it and it has been fine since then.
Cheers,
Jacqui
Jacqueline Ross
Biomedical Imaging Microscopist
Biomedical Imaging Research Unit
School of Medical Sciences
Faculty of Medical & Health Sciences
The University of Auckland
Private Bag 92019
Auckland, NEW ZEALAND
Tel: 64 9 373 7599 Ext 87438
Fax: 64 9 373 7484
http://www.health.auckland.ac.nz/biru/
-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On Behalf Of Nathalia Holtzman
Sent: 09 November 2007 02:59
To: [log in to unmask]
Subject: Z-drift
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi all,
I need some help. We have a 1 year old Leica SP5 inverted scope and pretty
much all of a sudden we started to get z-drift. Every time you do a time
lapse, no matter what the z-step or time interval, the system does not start
at the top of the sample, it moves down significantly and progressively.
Anything I can do to fix this?
I have called the service rep but would like to fix the problem ASAP. It is
possible that I am the first person to use the system for this application
since we upgraded the software and firmware.